<div dir="ltr"><br><br><div class="gmail_quote"><div dir="ltr" class="gmail_attr">---------- Forwarded message ---------<br>From: <strong class="gmail_sendername" dir="auto">Julius Madey</strong> <span dir="auto"><<a href="mailto:hillfox@fairpoint.net">hillfox@fairpoint.net</a>></span><br>Date: Thu, Jul 30, 2020 at 12:13 PM<br>Subject: new data<br>To:  <<a href="mailto:wittend@wwrinc.com">wittend@wwrinc.com</a>><br></div><br><br>
  

    
  
  <div>
    <font face="Arial">Dave,<br>
      This is the best data I have yet to demonstrate possible
      temperature effects on RM3100 measurements.  I had planned to
      stabilize temperature at near 0 C by mounting the 3100 adapter
      board on an aluminum block immersed in an ice/water mix, sitting
      in a foam box but a run this morning at 0500 to 0700 produced a
      variation of less than 0.5 C. <br>
      <br>
      The attached pdf includes a 2 hour run from yesterday night,
      2100-2300 edt, where temperature had not yet stabilized for the
      night but which , I think, clearly shows what appears to be
      consistent temperature related measurement effects. <br>
      <br>
      All measurements were made with a cycle count of 600 and a 0A
      register value of 10, which amounts to averaging 10 successive
      readings.  Status register query was held off until the reading
      was complete to avoid any possible effects of i2c bus activity. <br>
      <br>
      I am also sending this to PNI for their comment.<br>
      <br>
      Hope you don't mind using you for a post to the group.<br>
      <br>
      Jules<br>